Using Input Impedance to Measure Current
A project log for Low-noise, Easy-to-use Analog Data Logger: SiGZiG
Captures signals at the lowest noise levels with very low drift. Precise and easy to use. It’s all your data logging fantasies come true.
sigzigSigzig • 08/18/2015 at 05:59•0 Comments
Input impedance on measurement equipment is usually kept as high as it can in an effort to prevent loading down the signals the equipment is measuring. For example most general purpose digital multimeters aim for an input impedance of at least 1M ohm. The Zig-4 has a lower input impedance of 330k ohms but this is still high enough for many/most situations. That being said it is still important to understand input impedance and its affect on your measurements.
In this post I thought it would be fun to actually take advantage of the 330k input impedance of the Zig-4 as a sort of low current sensing.
The goal: To see if we can use the Zig-4 to detect a change in current in the very low nA range.
Even though most everything can be calculated, I thought it would also be nice to back up some of the calculations using actual measurements. I wanted to use a recently acquired Keithley 480 Picoammeter and on the lowest setting so the first order of business was generating a really small known current.
Using a 1G ohm resistor and a 1.25V voltage reference we can generate a current of 1.25nA.
The resistor is 1% and the voltage reference is not exactly 1.25V plus the meter could be off a bit... but overall that's close enough.
I then added a 100M ohm resistor in series with the 1G ohm resistor effectively lowering the current to a calculated 1.136nA
Again, close enough.
Now lets swap out the Keithley 480 picoammeter with the Zig-4 and do the same change in current to see what kind of signal we get. The current in both cases will be a little lower because now there is the additional 330k of the SiGZiG.
Above is the setup except I switched back and forth between the 1G and the 1G+100M ohm resistor. Below you can see the signal I was able to get. The Zig-4 does have a small amount of programmable gain, so in all fairness the below plot is with it set to a maximum gain of 8 giving it more bits for the range it was measuring.
So by using the input impedance of the Zig-4 to our advantage we can actually use it to monitor small changes in current down to the sub nA range. In the case above we were able to clearly pickup a change in current of only 115pA (113.5pA to be exact).